Because the v4.1 boards had possible noise issues, I've been looking a lot at quantifying what constitutes a normal amount of noise and what is "a hot mess". The ET-72/73 probes are pretty solid, the Thermoworks probes show more variation, and the ET-732 are pretty solid
unless something touches their braid.
EDIT: Information below is probably not accurate. See next post
What I've found is that the thermoworks and the ET-732 both show benefit from adding some capacitance to their output. It doesn't completely remove the variation but it cuts it in half for the thermoworks and reduces the chances of the ET-732 dropping off completely. The ET-732 dropoff is caused a 120Hz (2x mains frequency) small spike that lasts about 1/1000000th of a second. When HeaterMeter samples and gets this, it throws out the whole second's worth of readings causing the probe to drop out. Adding the extra capacitance attenuates this spike, although the extra noise caused by the ungrounded ET-732 probe shield can still cause dropouts at room temperatures. The only way to eliminate this is to solder a ground wire to the braid itself.
So it is win-win right? Well not 100%, the capacitance can cause an erroneous reading when plugging or unplugging probes. That's not so terrible though so here's something to try if your Thermoworks probes show a lot of variance or your ET-732s just drop in and out. This is for both v4.0 and v4.1 HeaterMeter boards.
Simply solder a 0.1uF capacitor from one side of the probe jack to the other. Use a ceramic / mlc capacitor such as
the same 0.1uF capacitor used in the rest of the build. The exact value isn't important but I've tested from 0.047uF (47nF) to 1.0uF and they all do alright. Even a low-ESR 10uF worked acceptably, but the chances of getting a bogus reading when plugging/unplugging go up the higher the capacitance.