try this for a test. plug the thermocouple into the board. measure between shield and one leg of the thermocouple on the board. It should read greater than 10 megs. The higher the better. Do this for the other lead to shield. measure across the thermocouple leads on the board. should measure 0 ohms. Last measurements will be between each thermocouple solder point and power ground. you should see 1k on each lead to power supply common. Do this again for shield. It should read greater than 10 megaohm. When it comes to noise, you need to make sure you have plenty of current when using a wall mounted power supply and also since we use switchers, the quality of the filtering can be very marginal. I always measure for ac ripple when I use wall plug power supplies. I first measure for ac ripple with my meter and if it looks high, then I look with scope. I do this open circuit and then with the supply connected to a load. I personally use only battery power with my heater meter. I do have a go to wall 3ADC supply that measures low for ripple and noise and will use that sometimes, but I mainly use battery power. C8,C9, and C11 must be correct values and have good solder joints. These components decouple the noise along with the 3 input resistors like Brian said. I am still suspicious of the thermocouple probe and power source. I use Thermoworks probes and have had no issues. Probes I have purchased off of e-bay have given me fits, mainly because of origin and they do not come with proof of performance. If you have a scope, repeat the same measurements and this time touch the shield while looking at the thermocouple inputs. If you see noise, I would find a new probe.