Recently, I've been making my own thermocouples, and today stumbled upon something odd. I make my thermocouples un-grounded, meaning the wire braid and metal sleeve are not connected to the thermocouple wire. I noticed, however, that noise would be induced on the TC when I touched the braid while barefoot, and also induce noise when hooked up to my WSM which sits on the ground (grounded). This occured with several different TCs I had.
Normally the thermocouple looks like this checking it with the hm-noise firmware.
When touching the wire braid (essentially attaching to house ground), it generates noise in the form of a cyclical (sinusoidal) wave.
So I thought, what would happen if I grounded the Braid to the thermocouple? I did, and the noise disappeared with no ill effect or deviation from the temp readings.
Now here's the strange thing. This noise only occurs with several different SWITCHING power supplies I had laying around. When I tested with a LINEAR supply, there were no ill effects grounded or un-grounded.
Should we all be using grounded thermocouples?
Normally the thermocouple looks like this checking it with the hm-noise firmware.
![ZF9R4ak.jpg](http://i.imgur.com/ZF9R4ak.jpg)
When touching the wire braid (essentially attaching to house ground), it generates noise in the form of a cyclical (sinusoidal) wave.
![F64VFM6.jpg](http://i.imgur.com/F64VFM6.jpg)
So I thought, what would happen if I grounded the Braid to the thermocouple? I did, and the noise disappeared with no ill effect or deviation from the temp readings.
Now here's the strange thing. This noise only occurs with several different SWITCHING power supplies I had laying around. When I tested with a LINEAR supply, there were no ill effects grounded or un-grounded.
Should we all be using grounded thermocouples?